If there is a way to remove the autosampler from the system and replace with a manual injector to perform full loop injection tests into the HPLC then one can isolate autosampler from the rest of the flow path.
If that is not possible, then (re)consider other sources within the autosampler. You reported that you have already replaced the syringe, checked visually for air bubbles and reconnected tubing/fittings.
Some other recommendations...(these are general autosampler recommendations):
* Injection port seal: has this been replaced? This is usually replaced during PM cycles and a worn injection port seal can often result in the symptoms you describe (lower peak area response, poor precision/injection linearity). Consult with your vendor's Operating Manual for the replacement operating cycle recommendation.
* Syringe: often there is a PTFE gasket in the syringe valve for the syringe port and in other syringe valve inlet/outlet ports that can be worn from repeated opening/tightening which will allow air entry into the injection path; if these have not been replaced then consider doing it (these are also replaced during PM cycles)
* Wash solution: as Tom indicated earlier, air bubbles are problematic; is your autosampler wash solution degassed before purging the transfer lines?
* Ausampler method parameters: even though you reported that your autosampler precision and linearity tests are poor, please double check your autosampler method parameters? Is the syringe volume correct? Injection volume OK? Injection method OK?
* Injection valve PM: You mention the injection valve as a possible source issue. You did not mention if you did the PM on the valve by rebuilding it by replacing the stator so this is another possible source if there is microchanneling from a worn valve.
If all the above check out then you will need to continue your troubleshooting downstream from the injection mechanism. I don't know if you are testing with a column inline...it would be best to replace the column with a backpressure loop as in an IQ test to eliminate the possible effects of the consumable (e.g., if there is a column capacity or other column performance related issue).
I agree with Tom that your absorbance detector performance (lamp energy, optical path issues?) may be the source of your problem. Please investigate the other recommendations first before going the hard route.
I hope these suggestions help...good luck!
very helpful advice , much appreciated. we had a waters technician come in and did some auto runs after filling the vials with water , the vials were still full even after the injection , in fact , the caps were wet which indicated that the injection valve wasn't diverting the mobile flow while it was making its injections thereby spilling mobile phase onto the top of the vial caps. He replaced the injection valve in its entirety which costs us 3000 dollars just for the parts , and then replaced the filter at the outlet of the tubing that leads from the syringe to the system flow. It worked for a day and now the problem is evident once again.
I've never actually opened up the injection valve myself , is a lot of technical expertise required to replace the stator on the injection valve?