unexpected area increase in hplc
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
What is the reason for unexpected area increase in hplc? Earlier was like 400000 and now increased to 700000 without any chanse in composition of mobile phase and column
-
- Posts: 325
- Joined: Fri Apr 08, 2011 6:15 am
Not reason:What is the reason for unexpected area increase in hplc? Earlier was like 400000 and now increased to 700000 without any chanse in composition of mobile phase and column
400000 units - its peak without liner scale of chromatograph . Light T% more 99 (D>2) in DAD , electonic signal is a biggest unit on analog-digital convenctor and other.
Who says the size of mountains, without seeing her top?
Also - in chromatography we see not absolute number of phisical parametrs , but the relationship these parameters portions between separated substances. Recalbrate system from standart solution and work...
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
What DSP007 is trying to say (I think) is that area counts are arbitrary (actually, the units would be something like mV-sec). That said, the first question I would ask is "are your retention times still the same? Next question: when you said "without any change in composition of mobile phase . . .", were you using the exact same batch of mobile phase and if so, how old was it? Sometimes seemingly insignificant changes (or aging) can change absorbance spectra significantly. That's why calibration should be done daily.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 2
- Joined: Tue Aug 02, 2011 2:17 am
I tried with same batch mobile phase and different lot mobile phase , everything gives same double area. second thing RT is same like previous , no change. Other thing, unit for area millivolts.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Autosampler settings? "Pilot error" preparing standards?
Can't be. Peak *height* would be in millivolts. Area units would have to be the product of Y-axis units (millivolts) and X-axis units (usually time).Other thing, unit for area millivolts.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts
Page 1 of 1
Who is online
In total there are 346 users online :: 3 registered, 0 hidden and 343 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Google Adsense [Bot], Semrush [Bot] and 343 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Google Adsense [Bot], Semrush [Bot] and 343 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science