Instrument method stop?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
7 posts
Page 1 of 1
I am having trouble setting up an instrument method and sample set. Currently, I have no problem with my instrument method when I do single injections. It starts, runs and stops as it should. However, when I set up a sample set, the method starts and runs the first injection, but then I receive an "instrument failure" notice and must "abort and continue on next line" with each new injection. The only way I've found around this is to remove the "0.0 flow" line from my instrument method. However, when I do this, the injections proceed with no problem, but at the end of the method, the pump does not shut down. At this point, I am unable to do runs overnight and such, which is the point of the autosampler! Any advice/help would be much appreciated!
-
- Posts: 276
- Joined: Tue Sep 07, 2004 10:05 pm
I dont what data system you are using but these are 2 things to try.
instead of zero flow, make it 0.01 mL/min
or
Make a new isocratic instrument method that has zero flow and run it as the last line in your sample set (i.e. a shutdown method).
AA
instead of zero flow, make it 0.01 mL/min
or
Make a new isocratic instrument method that has zero flow and run it as the last line in your sample set (i.e. a shutdown method).
AA
-
- Posts: 11
- Joined: Wed Jul 27, 2011 4:33 pm
I don't know if this changes anything, but we've got the following setup:
Waters 490E multiwavelength detector
Waters 510 Pumps
Waters 717 Autosampler
We're running Empower Pro software.
Thanks for the suggestions!
Waters 490E multiwavelength detector
Waters 510 Pumps
Waters 717 Autosampler
We're running Empower Pro software.
Thanks for the suggestions!
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
I think you can use your method as a gradient, with the last line a flow of 0.
But you should make the time of your last line long enough (say 5 minutes more than the normal runtime).
Only your last injection should run the entire gradient. (Or even not, your HPLC will continue the gradient if there isn't a further injection).
Ace
But you should make the time of your last line long enough (say 5 minutes more than the normal runtime).
Only your last injection should run the entire gradient. (Or even not, your HPLC will continue the gradient if there isn't a further injection).
Ace
-
- Posts: 2174
- Joined: Tue Aug 17, 2004 7:59 pm
^ same deal w/ detector events. You may be using one to shut off your lamp, make sure it doesn't shut off the lamp until ~5 min. after data collection ends for last sample.
Thanks,
DR

DR

-
- Posts: 62
- Joined: Wed Mar 02, 2005 9:39 am
One of the easiest ways is to create a STOP method. In the instrument method, you just use isocratic mode and use 0.00 ml/min as flow rate. You put this into a Method set.
The last line of your sample set, you should use "condition column" as function, used the STOP method as method to use, and use e.g. 1 min as run time. So at the end of your sample set, the stop method is exectuted. The method can also include the shut down of detectors etc.
I use this for year and had not trouble with it.
I hope this helps.
The last line of your sample set, you should use "condition column" as function, used the STOP method as method to use, and use e.g. 1 min as run time. So at the end of your sample set, the stop method is exectuted. The method can also include the shut down of detectors etc.
I use this for year and had not trouble with it.
I hope this helps.
Regards,
Gilbert Staepels
Ideas mentioned in this note represent my own and not necesseraly those of the company I work for.
Gilbert Staepels
Ideas mentioned in this note represent my own and not necesseraly those of the company I work for.
-
- Posts: 11
- Joined: Wed Jul 27, 2011 4:33 pm
So now that I've set up a stop method and switched my last method step to 0.01ml/min, I can no longer get the sample set to switch injections. I've tried generating a new instrument method and sample set with each set up, but to no avail. Is this just a bug with the software or am I missing something?
7 posts
Page 1 of 1
Who is online
In total there are 357 users online :: 1 registered, 0 hidden and 356 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 356 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 356 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science