Good suggestions by all.
1. The background trace overlays do look like 254 and 280 nm; the 280 nm (purple trace) has a lower profile than the 254 nm (black)---just hard to read the fuzzy numbers
2. I agree that the first thing to check is the water quality. I was not able to see any confirmation that HPLC grade (spectrophotometric) was used...we just need to confirm this. In any case, pop open a fresh bottle of HPLC grade water; or, if using a DI RO water system, have that checked, especially the TOC level for organic breakthrough.
3. Another consideration is the column set. If the column guard (or if no guard present, the separator column inlet) is saturated then running the gradient program on the blank will elute the retained material as the eluent strength increases from the gradient program. The column (and guard) can be cleaned according to the column vendor's instructions...this may take some time; replacing with a new column and after conditioning, then running the gradient will give you some helpful information right away (after verifying the water quality in #2 above).
4. If #2 and 3 check out, then back to square one. Hardware, perhaps. Running the system under hardware IQ conditions is the recommended approach; when presented with this type of situation, our approach is to verify the hardware functionality first, then add the chemistry next...
Hope this helps and good luck!