Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hey there. Just wondering if anyone else is doing any method work on this material. I'm trying to find a low cost column solution for a L1 packing 4.6 x 150mm column. I'm looking at a Symetry Shield from Waters but am looking for any other input that might run cheaper. Any help would be appreciated.
Hi,

I am not sure whether the USP monograph for Methylparaben is the same as in European Pharmacopoeia EP. Therein, also a C18 column is used (4.6 x 150 mm) using MeOH/Buffer 50:50 as mobile phase (buffer if I remember correctly: 6.8 g/L KH2PO4). In the EP knowledge database, it is outlined that a Phenomenex Prodigy is suitable or a GL Sciences Inertsil ODS-3. I tried Inertsil ODS-3 and it worked; and also Inertsil ODS-4 works.

Regards

Florian
Hi,

I am not sure whether the USP monograph for Methylparaben is the same as in European Pharmacopoeia EP. Therein, also a C18 column is used (4.6 x 150 mm) using MeOH/Buffer 50:50 as mobile phase (buffer if I remember correctly: 6.8 g/L KH2PO4). In the EP knowledge database, it is outlined that a Phenomenex Prodigy is suitable or a GL Sciences Inertsil ODS-3. I tried Inertsil ODS-3 and it worked; and also Inertsil ODS-4 works.

Regards

Florian
Thanks a bunch, I'll lokk into both of them!
Hello all.
LC chromatography and detect of methylparabene isn't problem. Its good solve in methanol (also in acetonitrile , alcohol , ect) small solve in water (1:80 (25C) -1:45 (90C)) , have hydroxybezene chromoforic group (ww1max=220nm E1%=200 ww2max=261 nm(pH1-8)E1%=30 or in Na salt (pH more 9)ww2max=290nm E1% about 70. pKa (phenyl) = about 3,5-4.
Simmetry, Dionex ( Nukleosil, Cromasil, Spherisorb others) - no particular difference, if L1 (C18) %C- more 5% and endcapping of silicagel.
Optimum organic phase 25-70% , optimum pH - smaller 3 or lager 5, optimum buffer molarity 0,05-0,5 M/L , optimum liquid phase speed - 1-1,5 ml (d=4,6 mm) or 0,1-0,15 (d=2mm), optimum T 20C to 40 C , need stabile temperature ( bad situation - after night in room 20 C air, sun peered through the window - in a room 30 C)

You problem -will be divided as the rest of the components of your sample. I know , that you sample - is a drugs (tablets, injections, liniments) or food. So they should check the completeness of extraction and the degree of separation the main components of the sample.
4 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry