What is the make up of your yearly sample load?
This will be for an R&D instrument to help figure out oddball stuff connected to an APGC interface to a QToF, as well as as a backup to the fixed loop HS-GC for FID assays (it also has an FID). I think it will be primarily for liquid injections without too much headspace workm (unless the fixed loop instrument goes down or we get a lot more samples than usual for that assay)
The fixed looped instrument does about 75-300 injections/week and occasionally more, with about 150/week being typical.
Since you've already stated that most of it is headspace, were you changing loop sizes often on your fixed loop apparatus?
The fixed loop instrument is always 1 mL and I'm doing a split injection with all the carrier flow routed through the headspace transfer line and split ratio of 5:1.
BTW this is a relatively new job. At my last company the GC work was exclusively headspace (various residual solvents from a lot of different matrices) and as i had multiple instruments I rarely has to change loops.
In this job I'm not doing residual solvents.
I'd be cautious regarding the savings in money. The combi-pal is like an inkjet printer. The basic models are essentially given away because they know they'll make a killing on the ink cartridges you'll need to buy.
Yikes!
If the majority of your injection sizes are between 500 and 2000 µl, you are in good shape, if you need to go outside that range then it gets expensive. Usually the system comes stock with 1 syringe heater meant for a 2.5 ml headspace syringes. However, there are 1ml and 5ml syringe sizes that need their own block heater. Those are pricey.
In the routine assay I'm doing I'm looking at high concentrations (.1% to 20% ) so I will be able to inject in that range as long as I can do a split.
For the other stuff I don't know yet.
Do you change temperatureof your vial heater and consequently your transfer line? Ideally, you would have dedicated syringes for different temperatures. For instance, USP <467> calls for 80°C vials. The syringe would need to be at 85°C. But if you have a vendor/in-house method that required a 105°C vial temp, then 110°C would be the recommended syringe temp. In this range of temperature, the teflon is significantly changed because it's malleable and would expand when heated. The chances for a bad seal would be highly likely if you went back to the original temps first mentioned.
How 'different' is different? is 10°C too different?
Sigh... looks like I made the wrong decision, as I don't know what I'll need other than for the routine assay. With the loop instrument one does not need to worry about such things. I suppose I can get and keep a few syringes for specific temperature ranges.
Be sure to have extra syringes available especially for the honeymoon phase with the instrument. There are no second chances with a zee'd needle.
OK.
Screw cap vials seem to be preferred over the standard crimp caps. May be more expensive.
Do you know why? Seems to me, it should not matter and that screw-cap would be more likely to leak as they could be sealed less reproducibly. For my fixed loop instruments, I aways use crimped vials because of that.
Other questions to ask yourself when choosing this system, how active are the compounds you intend analyze?
Although not the main targets, in the routine assay i do quantitate some short chain fatty acids... the levels are high enough that they are not a problem with the loop sampler.
Mainly because he's where I got the fraction to begin with.. For unknowns ... I don't know!
I had to do a trial by fire with the combi-pal and it was very trying, but I did gain a lot of knowledge.
Are you still using it?
BTW too I am very comfortable with the loop sampler as I have been using them for 16 years with relatively few problems and little maintenance beyond yearly PMs.
As I said I chose the Combi-PAL for flexibility, space savings and being able to use a single sampler.
Anything else, just let me know.
Thanks I appreciate it.
It will likely be several weeks before I start using it as it has not been installed yet and I am going on vacation soon.... but I'm sure I will have more questions once I get started using it!
- Karen