Advantages and disadvantages for ms-NoVent

Posted: Mon Mar 28, 2005 11:12 pm
by jimmyhu
The SGE sells ms-NoVent device that allows the MS to be operational almost immediately; results in minutes rather than in many hours or even days, depending on the sensitivity required from the instrument. It costs about US$1,000.

Is it widely used by GCMS analysts? What are the advantages and disadvantages for this device?

Any help would be greatly appreciated!

Jimmy

Posted: Tue Mar 29, 2005 2:45 pm
by Answer Man
Jimmy: I cannot help you with the SGE ms-NoVent, but I thought I would mention the Restek EZ No-Vent ($175 US), another common product for your application. I have attached a link which should explain the pro's and con's of this product. They work well, but they do allow a small volume of air into the MS (through a small 0.1mmID transferline). In addition, because of the transferline, your carrier gas pressure program will need to be modified (increased) to compensate for this restriction.

http://www.restek.com/fantasia/pdfCache ... 890-16.pdf

Posted: Tue Mar 29, 2005 6:05 pm
by Consumer Products Guy
I've had both in my lab and I like the Restek EZ No-Vent better, it's simpler as well. I did find the directions on "how much column end should protrude" a little unclear (but I'm not a rocket scientist), but it does work, and we don't get any air peaks due to leaking. We're using it now.