I had been running up to 100 injections per sequence (A few sequences every 2 weeks). Sample has been cleaned up using liq-liq extraction and SPE.
I am not sure if my samples are making the system dirty pretty fast or if there are other reasons. Our senior scientists insisted it is my melamine samples that "dirtied" the Q0 and Q1, leading to frequent cleaning of the system.
Before we had the problem of melamine polymerizing on the spray tip. I inserted blanks between each set of samples and adjusted the position of the spray tip to avoid this problem.
My LC conditions are:
(A) 10mM ammonium acetate
(B) ACN
Gradient of 95% (B) to 50% (B) from 0 to 6 min, 95% (B) from 6.1 to 10 min.
Flow rate of 0.25mL/min
I used the diverter valve to direct eluate to waste at 0 to 3 min and from 6 min onward. I had also used a guard column with my Atlantis HILIC column as well.
Any advice or comment would be appreciated. Thanks!
-----