Posted: Tue Jul 26, 2011 7:42 am
by zaidalisharif1
hi
all
I am working with aglinet cp 3800 i am getting low nitrogen responce please can i know the reasons for low nitrogen responce
Re: low nitrgen responce
Posted: Tue Jul 26, 2011 10:06 am
by chromatographer1
A low concentration of nitrogen could be the cause for the low response, but that is obvious. What is not obvious is what you are doing, how you are doing it, and how much you think your result is in error.
Can you be more forthcoming?
best wishes,
Rod
Re: low nitrgen responce
Posted: Tue Jul 26, 2011 12:59 pm
by zaidalisharif1
the concentration of n2 is given on certificate is .20 but i am getting .197 ...it is the calibration standerd
when i run other standerd it is ok with diff concentration standerd .... it is happenig only with new standard
Re: low nitrgen responce
Posted: Tue Jul 26, 2011 1:46 pm
by AICMM
zaidalisharif1,
Odds are good that your blend tolerence is 2% and your value is within that window. Perhaps your blend is just a bit low.
Best regards,
AICMM