AICMM » Tue Jul 26, 2011 1:27 am
DaBrainz,
First, the direct answer to your question. Yes, a TCD is swappable with an FID if you have the mechanical abilities and the parts. It's tedious but not rocket science.
Second, I am with Chromatographer1 in saying that you should think about an HID. I have a well known bias in this regard in that I make an HID but there are other reasons as well. For one thing, if you are looking at low levels of O2, the HID will be much, much more sensitive. Furthermore, it will not be all that difficult to install an HID in place of a TCD on a 5890 chassis (I have done this numerous times) since you can use the same electronics, plumbing and heater with minor modifications. Finally, HID's are not really all that fussy as evidenced by numerous customers of mine that have them installed in various GC's, process and laboratory. I don't use a getter (nor do most of my customers) and I don't use any fancy plumbing and I can easily see 10 ppm or less of the fixed gases. Considering you already have the FID electronics, plumbing and heater, this will also be a relatively inexpensive option, especially considering how hard I think you will have to look for a 5890 TCD (unless you already have one!)
I believe the Valco PDD will retro-fit on a 5890 without too much work, I know mine will, and I am almost certain that a Gow-Mac DID will not since it is normally part of their own GC.
If you would like me to talk your ear off about this, feel free to contact me at aicmm at flash.net
Best regards,
AICMM