Connection problem!Empower with Agilent 1260.

Posted: Fri Jul 22, 2011 6:43 am
by mars2011
Hello,guys

Maybe someone could help me on that issue on Empower 2 :

Agilent 1260.

Can't connect,just "instrument Failure".

What happened?

the connection is very well,when that system work with 1200

Re: Connection problem!Empower with Agilent 1260.

Posted: Fri Jul 22, 2011 11:07 am
by unmgvar
make sure you have the correct empower release feature
check out the firmwares on the system
are they the minimum requirement or above from software specs

Re: Connection problem!Empower with Agilent 1260.

Posted: Fri Jul 22, 2011 9:22 pm
by Bull76
Hi,

The 1260 isn't as of yet supported by empower. You do need the latest ICS 1.06 at a minimum for the control of Agilent equipment. Also what level of empower are you?

It can be connected and used in empower.