Saturation Columns - do they work?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

9 posts Page 1 of 1
We run a method using a silica column and pH7.2 mobile phase. A certain school of thought suggests that a saturator column retards the errosion of the analytical column packing. Another school of thought suggests that this is a waste of time.
Wisdom please.

JW

Guard columns can work in "sacraficial mode" as long as you remember to replace them before they're gone. They don't eliminate erosion, but they can retard it some (so I am told).
Thanks,
DR
Image

It is not impossible that the solubility of the silica limits your column life, and that this could be fixed by using a saturator column.

On the other hand, the problem of column life could also be due to contamination. In this case, a guard column would serve you best. A guard column actually will have two functions: it will protect the column from dissolution and from contamination. I would go for the guard column approach.

I have tried the saturator ("precolumn") trick a couple of times -- admittedly years ago. It didn't seem to make any difference.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Thanks for the answers.
Has anyone thought of pre-saturating MP w/ fine silica particles during prep? Filtering afterwards of course.
John

Not that I'm aware of. I'd worry about that on the chance that the solubility of the silica might decrease (e.g., due to a temperature change), resulting in precipitation of silica where you don't want it. (Hey, just because I'm paranoid doesn't mean that someone's not out to get me! :wink: ).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Don´t you use glass bottles?
(Is this business of protecting main column dissolution with a guard one of the myths of chrom?)

Yes we do use glass bottles. Good point.
Glass is not the same as column silica, however, and doesn't have as much surface area.

You're probably right in that this is a myth. No one has said it has definitely improved the life of their columns.

John

But consider the time the stuff spends in the glass bottle vs the time in the guard. On the chemical difference? Glass is certainly "dirtier", probably dissolves more easily??
9 posts Page 1 of 1

Who is online

In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry