sulphonated polymer retantion on SAX column

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi all,

we have a method which we use ammonium acetate as the mobile phase and use sulphonated polymer diluted in water through a sax column. We are using UV detection. We seem to be retaining the sulphonated polymer on the column and are getting no peaks. Anyone have any idea why or indeed how we can get it off the column without damaging it?

x
You may not be able to get it off the column. :cry:

Things to try:

- increase the concentration of ammonium acetate
- add a salt of a polyvalent anion t (e.g., sulfate) to the mobile phase
- try adding some organic solvent to the mobile phase (if your polymer is only lightly sulfonated, retention may be partially due to hydrophobicity.

For future use, a weak anion exchanger may be a better choice, because that will let you manipulate the charge on the stationary phase by varying the pH.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 52 users online :: 2 registered, 0 hidden and 50 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 50 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry