Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
hi
all
I am working with aglinet cp 3800 i am getting low nitrogen responce please can i know the reasons for low nitrogen responce
A low concentration of nitrogen could be the cause for the low response, but that is obvious. What is not obvious is what you are doing, how you are doing it, and how much you think your result is in error.

Can you be more forthcoming?

best wishes,

Rod
the concentration of n2 is given on certificate is .20 but i am getting .197 ...it is the calibration standerd
when i run other standerd it is ok with diff concentration standerd .... it is happenig only with new standard
zaidalisharif1,

Odds are good that your blend tolerence is 2% and your value is within that window. Perhaps your blend is just a bit low.

Best regards,

AICMM
4 posts Page 1 of 1

Who is online

In total there are 28 users online :: 0 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry