Advertisement

Reasons for >100% extraction recoveries

Discussions about sample preparation: extraction, cleanup, derivatization, etc.

4 posts Page 1 of 1
Given that extraction recovery values had been corrected for blank values, what are the possible reasons for extraction recoveries of more than 100%?

If recovery values are between 95% to 105%, they are typically accepted. Is that extra 5% used to account for random errors?
It could be just the variability in your whole analysis. I know for GC/MS methods to expect less than 5% variability for the entire methods process (including sample prep etc.) is kind of unusual. Positive bias on a GC system could be related to active sites in the chromatography system. Example: Sometimes with new GC liners and columns and for certain analytes there are active sites which cause the system to give a lower response than they will after running a few samples. i.e., If you calibrate on a new liner with new column, the samples or standards you analyze after this may be biased slightly high. I have seen this happen mostly with GC/MS analysis of PAHs and also on GC/ECD for PCBs. I would not be too concerned about a 105% recovery. If you look at U.S. EPA methods http://www.accustandard.com/asi/pdfs/ep ... /8270D.pdf you will see many tables with extraction stats for many semivolatile compounds. Note that some % recoveries are greater than 100%. This is acceptable for some compounds. If you're working with LC then I'm far from an expert but I would imagine that a similar trend would be observed as it is still related to extraction and chromatographic principles (ab/adsorption etc.)

Hope this helps.
~Ty~
Is that extra 5% used to account for random errors?
As a crude approximation, yes. It's outside the scope of a Forum post, so grab a statistics textbook and read up on "confidence intervals".
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
tlahren: Thank you so much. :) That helps a lot in answering my questions.

tom: Noted. Thanks.
4 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry