FID hydrogen problem

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
Hello coleagues
I use shimadzu GC-17A v3 with Class VP 4.3 software and ss420x A/D bord.
I have problems with H2 flow of FID. When I download a method's parameters to the instrument hydrogen flow turns off. to obtain chormatogram I have to turn H2 on by hand and to reignite detector while autoinjector prepares to inject.
This happens since I changed the computer of the system. I installed and configured the software according to instructions and the same parameters of method I used when everything was ok. The problem occures on Win XP and VISTA, I tried both OS but this isn't the reason.
Gas pressure is within the limits, FID is turn ON both on GC and software. If I start a run from autoinjector by hand everything is ok, but when I start it from software- H2 turns off.
Please give some advises what to do...
thanks and greetings

Achkata
Hello Ashkata
1) Water sample - water vapor turn flame off , not flame signal - hydrogen valves line off.
2) Also - valves off , or pipline (tube) or nuts destroy (not germetical)
Use foam measuring the velocity ( г (Russian G in manual wrote as 2 in picture) of the gas flow in order to verify that the hydrogen enters the detector in the right quantity.
Image
Could there be a command inserted in to the method that is shutting the detector off?
hi
thanks for replies...
the problem definitely is with the software (ClassVP). After few days of boring fight, I repaired old computer, there was a hardware problem, and I have working system now. Having two computers with Class VP-working and not working, I compared all settings, adjustments and parameters that I know and everything is the same...So I still looking for this command that turns the detector off since the begining of the problem. There is a check box for detector on/off and it's on.
Shimadzu Representative was surprised that the system works with WIN XP because ClassVP4.3 is too old. He supposed that a conflict occures between ClassVP and WIN XP.
I gave up for now because 'good old computer' works. In case of future computer problems I'll have to find where the problem is or to buy new software.
Meanwhile I accept advices.
thanks and greetings
Try copying the software from the old computer to the new one, or just copy the files containing the settings.
5 posts Page 1 of 1

Who is online

In total there are 28 users online :: 0 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry