Hi Guys,

I am suffered a lot from the placebo interference from my syrup sample. I could not resolve my impurity peaks from placebo peaks. I tried with different columns and mobile phase. My sample has 2 actives, methylparaben, propyl paraben, sorbitol, FDC red40, Sugar, Sodium Citrate and citric acid. Due to one of my active components nature i cannot go for more sample concentration and higher wavelength. My working wavelength is 205 nm. Anyone can help me to resolve this.

thanks in advance

Re: Placebo Interference

Posted: Mon Jul 18, 2011 9:20 am
by Gerhard Kratz
Maybe you should get rid of the matrix effect by doing SPE sample preparation. Did you also try a HILIC column?

Re: Placebo Interference

Posted: Mon Jul 18, 2011 11:04 am
by sank_yb81
Hi Gerhard,

Can you explain how to perform SPE. Because this is the first time i am going to perform.

HILIC column will work? I will try it today and get back to you

Re: Placebo Interference

Posted: Tue Jul 19, 2011 10:57 am
by sank_yb81
HILIC Column also not good enough to resolve the peaks

Re: Placebo Interference

Posted: Wed Jul 20, 2011 8:01 am
by Gerhard Kratz
I would do use a C18 SPE to clean the sample, using an Water as solvent. Which HILIC column you tried?

Re: Placebo Interference

Posted: Wed Jul 20, 2011 10:50 am
by sank_yb81
Hi Gerhard,


I used waters atlandis silica HILIC column