I'm dealing with HPLC to analyze gamma aminobutyric acid (GABA) with the following method:
-Sample pretreatment: 100ul GABA + 100ul 0.2 M carbonate buffer pH 9.8 + 200ul Dansyl chloride (8g/l) --> mix and heat block at 80 degree for 40 min---> add 8ul Acetic acid (10%).---> mix then centrifuge (12.000 rpm for 5 min)
-Mobile phase: Eluant A: THF/ Methanol/50mM sodium acetate pH 6.3 (5/75/420)
Eluant B: Methanol
-Column: OptimaPak C18 (4.6 X 150mm)
-Elution conditions: Equilibration 6 min (20%, B)
Gradient 20min (20-80%, B)
Cleaning 3 min (100%, B)
flow rate: 1ml/min
measurement : UV 286nm
But there is one thing happen.
With small concentration of GABA I got the higher peak than the higher concentration.
Have you ever face this problem before?
Can anyone explain for me about this. Why did it happen and how to solve the problem?
Sometime I obtain peak, but sometime doesn't, why?
Please help me
(
Thank you so much.....
-Sample pretreatment: 100ul GABA + 100ul 0.2 M carbonate buffer pH 9.8 + 200ul Dansyl chloride (8g/l) --> mix and heat block at 80 degree for 40 min---> add 8ul Acetic acid (10%).---> mix then centrifuge (12.000 rpm for 5 min)
-Mobile phase: Eluant A: THF/ Methanol/50mM sodium acetate pH 6.3 (5/75/420)
Eluant B: Methanol
-Column: OptimaPak C18 (4.6 X 150mm)
-Elution conditions: Equilibration 6 min (20%, B)
Gradient 20min (20-80%, B)
Cleaning 3 min (100%, B)
flow rate: 1ml/min
measurement : UV 286nm
But there is one thing happen.
With small concentration of GABA I got the higher peak than the higher concentration.
Have you ever face this problem before?
Can anyone explain for me about this. Why did it happen and how to solve the problem?
Sometime I obtain peak, but sometime doesn't, why?
Please help me
Thank you so much.....