WATERS 2475 Fluorescence detector problem
Posted: Wed Jul 06, 2011 4:56 pm
After powering up, shows " Electronic A/D failure" and "Reference A/D saturated "
Any ideas ?
Thanks,
Tony
Any ideas ?
Thanks,
Tony
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://www.chromforum.org/
usually front amplifier PCB issue. How old is the detector and has anything been done to it, i.e a service or PM?