Hi all

I am working on a related compounds method using xterra column RP8 3.5 um : MP is ammonium bicarbonate buffer10mM pH10 and ACN gradient. When i condition the column i see an interference peak comes at 50:50 buffer: ACN; This interferes with my impurites. i couldn't get rid of the peak. It is present in water. i used high purity water. it is still there. If i do only 50:50 isocratic i dont see that peak. my gradient starts 95:5(buffer:ACN) stays for 5 mins. goes to 50:50 in 20 mins. i need this gradient for separation of 7 impurities in the method.

can some body help me with this ( i don't want to develop a method around an interference peak)

Wavelength used is 230; present is all wavelengths


thanks

From your post, I infer that you see this interference even when you run a "dummy" gradient (i.e, either no injection, or injection of mobile phase).

Try running two such dummy gradients following different equilibration time; for example:
- dummy gradient to flush the system
- equilibrate for 10 minutes
- dummy gradient #1
- equilibrate for 30 minutes
- dummy gradient #2.

If the interference peak is larger in gradient #2, then you have confirmed that you are looking at contamination in your A solvent. Possible sources:
- the water supply (obviously); try a different source of water.
- your buffer salts (try different lots and see if that helps)
- the pH electrode (if you are dipping the electrode in the solvent, you may have buffer leakage; try measuring pH on aliquots removed from the main vessel)
- contaminated glassware (detergent?)

When you solve the problem, let us know what you found (so we can all learn!)
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Majestic-12 [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry