You have not indicated a matrix at which you are looking at CO2 but any porous solid phase support can be contaminated and if the pores are partially blocked, the result is exactly what you have described.
The solutions to CO2 analysis by Gc are rather limited. Almost all use gas solid chromatography to perform the separation.
If oxygen and nitrogen are to be separated, as well as CO2, the limitations are even greater.
Molecular Sieve 13X will perform the analysis at room temperature or even higher (70°C) temperatures.
Using Carboxen 1000 and ambient temperatures a 15ft to 30ft 1/8 inch OD column will perform nicely. A temperature oven ramp is required for a shorter time analysis.
Alternately, a Carboxen 1004 1/16 inch OD column, 3 to 9 meters long will provide H2/O2/N2/CO/methane/CO2 separation and not be affected by water in your sample.
These are available from my company, Supelco.