both negative and positive peak in my chrom

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
I am doing method development for chloride leucine and lactate. the column is metro sep anion dual 2. My condition is 4mM NaHCO3 and 0.4mM Na2CO3 without suppressor. Right now the leucine and lactate peak are positive but choride peak is negative, how can I make them all positive?
Actually, the cloride peak is positive and the other two are negative. Chloride has a higher equivalent conductance than does bicarbonate, while leucine and lactate have lower equivalent conductances. In ion chromatography, the polarity of the signal is flipped as necessary to make the peaks go in the "proper" direction.

The only thing you can do with this system is to reverse the polarity while the chloride peak is eluting.

The alternative is re-develop a system using a buffer system whose equivalent conductance is lower than those of leucine and lactate (p-hydroxybenzoate, for example). Then all of the peaks will go the same way.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
thank you very much, Tom. Is suppressor will make any difference, since the paper I read with the similar condition, their peaks are all positive with suppressor.
Yes, the suppressor will make a difference. It replaces the sodium counterions with hydronium. That converts bicarbonate to carbonic acid, which has a very low equivalent conductance.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Maybe I should buy a suppressor, is that expensive? Our HPLC is waters, where can I find one?
Thank you very much.
Dionex, Metrohm, or Grace/Alltech would be three possibilities.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Without having experience analysing leucine with IC, I doubt whether you will see leucine sensitively after suppression as it will be converted into the zwitterionic form which most probably shows only little conductivity. Therefore non-suppressed might be the better solution.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
7 posts Page 1 of 1

Who is online

In total there are 47 users online :: 0 registered, 0 hidden and 47 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 47 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry