Help wrt. separation needed !

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
I am trying to remove these product related impurities that pops up as pre and post peaks around our product on our analytical system. The only preparative techinque that works in lab is a weak anion exchanger, but the impurites are comming out as shoulders effecting the purity.

I have tried playing with the temperature, pH and conductivity gradient. Any idea how i can have shoulders to come out as peaks ?

BR
Possibly:
- decrease the load
- use a smaller particle size (if you are not overloaded!)
Certainly:
- use a longer column (and a proportionally longer gradient)
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
can you provide more information on the nature of your compound?
you talk of using an analytical system, yet you say you do purification

what are the details of the analytical method you have and how different is it from the prep method you are using?
maybe you do not have all the parameters transferred correctly between the 2 columns sizes
in most cases when prep is done in an analytical system, flow rate are not used matching the linear velocity between the 2 columns leading to differences in the chromatography and purity
Possibly:
- decrease the load
- use a smaller particle size (if you are not overloaded!)
Certainly:
- use a longer column (and a proportionally longer gradient)
Load is already decresed to the absolute minimum.
We are using a very small particle size ( 15 microns )
The lenght of the column is 20 cm and any thing longer that that gives us pressure problems once we scale up !
can you provide more information on the nature of your compound?
you talk of using an analytical system, yet you say you do purification

what are the details of the analytical method you have and how different is it from the prep method you are using?
maybe you do not have all the parameters transferred correctly between the 2 columns sizes
in most cases when prep is done in an analytical system, flow rate are not used matching the linear velocity between the 2 columns leading to differences in the chromatography and purity
Sure , its not a compund but instead a protein ( 15 Kda)

We use a C18 (RP) and acetonotril in our analytical system which is something we cannot use once we scale up.
In preparative mode we have a strong anion exchanger with non organic buffers. But I am having problems getting the very close product related impurities we see on analytical system to separate on the preparative system !
is it an AKTA system?
what is the resin exactly you are using and mobile phase?
what are the impurities nature that made you choose the WAX?
are we talking of mono, dimer, HMW or is it something else?
moe like:
15 kda is something like lyzozyme and I have know an application for it on SCX and not WAX. maybe it would be better for you?
6 posts Page 1 of 1

Who is online

In total there are 46 users online :: 0 registered, 0 hidden and 46 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 46 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry