I have tried playing with the temperature, pH and conductivity gradient. Any idea how i can have shoulders to come out as peaks ?
BR
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Load is already decresed to the absolute minimum.Possibly:
- decrease the load
- use a smaller particle size (if you are not overloaded!)
Certainly:
- use a longer column (and a proportionally longer gradient)
Sure , its not a compund but instead a protein ( 15 Kda)can you provide more information on the nature of your compound?
you talk of using an analytical system, yet you say you do purification
what are the details of the analytical method you have and how different is it from the prep method you are using?
maybe you do not have all the parameters transferred correctly between the 2 columns sizes
in most cases when prep is done in an analytical system, flow rate are not used matching the linear velocity between the 2 columns leading to differences in the chromatography and purity