random nosie in FID

Posted: Fri Jul 01, 2011 4:49 am
by mr3573
What could be reason of random negative-positive noise in FID.? I Conditioned 3 hours and also change flame tip and tube collector. But no useful
Image

Re: random nosie in FID

Posted: Fri Jul 01, 2011 5:16 am
by Axar
Hi!
What gas flows do you set in FID?
Does your gas generators in proper condition?

Re: random nosie in FID

Posted: Fri Jul 01, 2011 8:35 am
by mr3573
Hi!
What gas flows do you set in FID?
Does your gas generators in proper condition?
the GC is three channel and other channel are ok. anyway carrier gas is heluim, makup is nitrogen from plant, hydrogen from generator and air from plant. if problem was related to gas we should have some noise in other channel but only one have noise!

Re: random nosie in FID

Posted: Fri Jul 01, 2011 5:00 pm
by chromatographer1
Could just be a detector that needs a regular cleaning maintenance.

Rod

Re: random nosie in FID

Posted: Fri Jul 01, 2011 6:20 pm
by krickos
mm

might not be related, but I occassionnaly find it suprising, that when operators change flame tip, they do not spend 30s to check for ferrule residues (from overtightning etc), that may cause similar random spikes, depending on desing I suppose.

Re: random nosie in FID

Posted: Sat Jul 23, 2011 7:08 am
by mr3573
finally found the problem. that was related to electronic FID board. thanks for all replyes