Hello everybody,

Is it possible to use different solvent for calibration standards (for example DCM or iso-octane) than for the samples (for example in hexane)? Can different solvents cause shifts in retention times?

Can we inject dichloromethane (DCM) with micro-ECD with appropriate solvent delay?

I will appreciate your input regarding these very basic questions. :)

regards,
Words are so important:

"Can we inject dichloromethane (DCM) with micro-ECD with appropriate solvent delay? "

Yes.

but SHOULD you? With a packed column I would say NO. With a capillary, I would say yes.

I would use the same solvent for samples and standards, why differ?

"Can different solvents cause shifts in retention times?"

It is possible, but not likely.

best wishes,

Rod
Words are so important:

"Can we inject dichloromethane (DCM) with micro-ECD with appropriate solvent delay? "

Yes.

but SHOULD you? With a packed column I would say NO. With a capillary, I would say yes.

I would use the same solvent for samples and standards, why differ?

"Can different solvents cause shifts in retention times?"

It is possible, but not likely.

best wishes,

Rod
Hi Rod,

Thank you so much for the reply. Yes, I have asked that in context of capillary column chromatography. Sorry for not being so explicit. The reason for using different solvents is in fact to reduce some steps in term of solvent switching. . Some of solvents are not compatible with the PDMS chips which I am using for the sorption studies of hydrophobic compounds. Making stock solutions of PCB and PAHs powder in DCM can make life easy because later on I can use the same to spike the methanol which i am using to load PCBs and PAHS on these PDMS chips. For the back extractions of these chips, I will use hexane/iso-octane.

regards,
Have you considered using methyl formate or ethyl acetate instead of DCM?

best wishes,

Rod
Changes in retention times with different solvents are unlikely, but changes in peak area are quite likely, depending on the inlet and injection conditions.

Peter
Peter Apps
Thanks Rod and Peter.

Yes, I can even use ethyl acetate but from Peter's reply it seems it is important that I use the same solvents for calibration standards and sample standards. I will be using typical splitless mode with purge valve open after 60 sec and temperature 270 degree C and injection volume of 1 microliter.

regards
If you do use different solvents you will have to run a series of comparisons between solvents to demonstrate what, if any,difference the solvent makes. That extra work may be more trouble than the solvent exchanges.

Peter
Peter Apps
Peter,

Yes, I see your point. Thanks for your valuable input.

regards
8 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry