connecting two or more columns

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
I connected three columns by universal connector as follow:
1- CPWAX 60M, 0.32
2-CPSIL 8CB 30M, 0.25
3-CPSIL 5CB 60M, 0.25
I needed low temperature for best separation. So first peak appear at about 31’ and peaks not so sharp (the peaks heights are short). Is it related to connector and if I use one long column have same situation or not
You have 150m of column, with two different diameters and two connectors so I would not expect the peaks to be very sharp. Using a single column can only improve matters - are 150m columns available ? Maybe you shouuld think of using a thicker film of stationary phase. What is your inlet pressure, and the flow rate of the carrier gas ?

Peter
Peter Apps
You have 150m of column, with two different diameters and two connectors so I would not expect the peaks to be very sharp. Using a single column can only improve matters - are 150m columns available ? Maybe you shouuld think of using a thicker film of stationary phase. What is your inlet pressure, and the flow rate of the carrier gas ?

Peter
The constant flow is 1 ml/min with initial pressure 18 psi. Actually I want to use a 100m column in replace of those three columns. But not sure to purchase new column.
What are you trying to analyse ?

Peter
Peter Apps
What are you trying to analyse ?

Peter
The analysis is for possibility of some individual diens in benzene. Desired compounds and their boiling points as per MSDS are:
A:2,4 Hexadiene(mixtures of three isomer);82 C
B:1,3 Cyclohexadiene;80 C
C:1,4 Cyclohexadiene;88 C
The original method used column number 1 and 2 in series and connected by usual glass connector. In that case all three compounds appear before benzene but one isomers of A co-eluted with non aromatics presents in benzene and therefore difficult to determination.
I tried to using some combination of columns
1) One polar followed by one non polar (column 1 and 3) not only couldn’t separate isomers’ A but also component C disappeared and I noticed C completely covered by benzene big peak
2)one or two of non polar column – although column number 2 is some polar!-(i.e. column 3 or columns 3 and 2 in series) haven’t been successful.
Best combination was by three columns but in that case height of peak have not been good (25-35 uV for 10 ppm standards)
Some time age I posted question about that in forum but we have not heart cut or switch as recommended in forum.
Sorry for long explanation and not so good English!
I would have advised heart cutting, but you say that you cannot do that.

Rather than trying to do this by the brute force approach of longer and longer columns perhaps a better approach would be to try to find a particular stationary phase that has the selectivity that you need. Try asking the column manufacturers for advice.

Peter
Peter Apps
6 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry