Nexera uHPLC (LC-30AD)

Posted: Tue Jun 28, 2011 6:50 am
by jusuom
We just recently purchased a Nexera UHPLC system (binary). Any users of the system here? What are your experiences? What to take into account when using the system? Have you pumped hexane sith Nexera?

Re: Nexera uHPLC (LC-30AD)

Posted: Tue Jul 19, 2011 2:22 pm
by Rob Burgess
I'd be interested to know how you get on with baseline mixing noise. We had a look at this syetem and found unacceptable what appears to be excessive mixing noise running water/acetonitrile gradients with TFA at low wavelengths. I know this is the extreme test for any systems but it just seemed to be worse on the Nexera than it did on an Acquity or 1290.

Re: Nexera uHPLC (LC-30AD)

Posted: Tue Sep 20, 2011 1:52 pm
by ginebra
if the baseline is not good , you just need to use either the 180µL mixer or both of them in line 180µL + 20µL and this will improve the basleine

Re: Nexera uHPLC (LC-30AD)

Posted: Tue Sep 20, 2011 2:21 pm
by Rob Burgess
We used both of those mixers and had special ones made for us - still no joy :(

Re: Nexera uHPLC (LC-30AD)

Posted: Tue Sep 20, 2011 8:12 pm
by unmgvar
with all UHPLCs systems you must tell your vendor that you mean to use Hexane
no UHPLc comes ready to use for hexane in long period of time.
in some cases you can use the system for a few days to 2 weeks, but then you must wash with IPA and reverse back to other solvents, otherwise you will kill the pump seals and other consumables
also you need to see that the waste tubing can deal with organics like hexane
it is not trivial at all

Re: Nexera uHPLC (LC-30AD)

Posted: Wed Sep 21, 2011 8:34 am
by Rob Burgess
Not sure why you are mentioning hexane as a solvent - we never used it on the nexera system we had on loan. We stuck to standard reversed phased solvents i.e. water & acetontrile + 0.1% TFA.