If you have cleaned your detector and the signal is still HIGH then you may have contamination in one of your gas supplies.
Yah! I also suspect that.
The air and hydrogen must be clean. Since the signal varies (but otherwise the detector seems to be working normally?) that means your contamination is varying. Are you sure it is not column bleed?
Change out cylinders or put new traps on your gas supplies. Is the signal 'noisy' or is the output smooth?
I'll try to find new traps and report you results. I used different capillary colums. Most of them was new and hadn't shown such bleed at other GCs. Signal output is hight, flucluating in a long term baceline. The word "noisy" is not correct - my fail. It is smooth enouth to recognise peaks of analites.
It sounds like you have someone with experience look at the GC, whether it is a service person or a chromatographer from other group.
There are only two persons, experiensed with GC in our organisation - me and my supervisor. The servise person could not help us.
You could be having an electronic problem as well.
It is the worst option, so I wish to don't think about it right now.
Too bad you can't show us a chromatogram expanded to show the noise level and the output level.