Small i.d. columns have been around in gas chromatography for a long time. In LC, I believe it was Endele who worked with a very small ratio of column diamter to particle diameter first, around 1972/1973, but the fact that at these ratios the structure is looser must have been around already before that, i.e. from GC.
Marc is correct: the rule of thumb is that once you get to a column to particle diameter ratio over about 30, the thing is a standard packed bed, and there is no difference in column performance, if the column is packed properly. Of course, for columns with a low aspect ratio (L/d), one runs into a lack of radial exchange, which can cause problems if the column packing is non-uniform. Giddings talked about the interchange of various non-uniformities in his coupling theory in the Dynamics of Chromatography from 1966.
Bottom line: a properly packed column works the same way, once the column to particle diameter becomes large enough.
The A-term of the van Deemter equation represents the non-uniformities in the packed bed. This includes the lack of uniformity between the wall and the center of the bed. Interestingly, Golay mentioned at some point that in an ideally constructed packed bed, this non-uniformity term should go away. I agree with this, because we reduced the non-uniformity term to about half a particl diameter with radially compressed columns. (For reference, a decent uncompressed column has a van-Deemter A-term of about 2.)
In electrokinetic chromatography, the fluid is transported forward by movement along the surface of the particles (or the wall in open tubes). If you visualize the open tube as an example, there is no frition between the different fluid layers, and thus the flow is a plug flow. In pressure-driven flow in an open tube on the other hand, the velocity is 0 at the wall and maximum far away from the wall. This difference is the basic reason for the higher performance of electrochromatography.