I recently started using a passivation technique on Agilent 1100 and 1200 series hplcs. I have checked with my Agilent rep and they gave me a recommended procedure also stressing the importance of removing the degasser and detectors from the flow path (and obviously using a column blank rather than a column). The procedure I am using goes something like this:
1. Purge at 5mL/min with water for 5 mins on both channels (binary pump)
2. flush system with water at .5mL/min for 30 mins (so far just to clean the system)
3. purge both lines at 5mL/min for 5 mins using 20% nitric acid
4. Flush system with 20% Nitric acid and .5mL/min for 30 mins
5. Repeat water rinse (steps 1 & 2)
6. Purge with 12% ammonia, flush system with 12% ammonia
7. Repeat water rinse (steps 1 & 2)
8. Purge with 1% formic acid, flush system with 1% formic acid
9. Purge with water, leave system flushing with water at 0.2mL/min for around 48 hours, until pH of effluent water is neutral.
I've carried out this procedure several times with no hiccups however recently a colleague ran a method with a triethylamine mobile phase (specifically pH 2.5 triethylamine buffer as the aqueous mobile phase and acetonitrile as organic mobile phase). On injecting a blank/diluent a large peak is eluting just after the injection peak. One theory is that the triethylamine is reacting with the oxide film provided by the passivation technique and is being released on injection.
I'm not quite sure how to troubleshoot this, so far I've flushed the system with organic solvent and cleaned the injector both with buffer and organic. Have also changed the PTFE frit at the purge valve and the glass in-line filters were not used during the passivation.