Need LC/MS compatible mobile phases with large pH range

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Looking to get some information from some of you experts out here. I am needing to set up some mobile phases from pH 2 to 11. I am thinking of using a 10 mM Ammonium Formate aqueous solution and using Formic Acid to generate two acidic mobile phases of pH 2 and 4.5 and using Ammonium Hydroxide to generated two basic mobile phases with pH 9 and 11. Utlimately I am needing about 5 mobile phases having pH's of 2, 4.5, 7, 9 and 11.

Does anyone see a problem with this from a LCMS perspective? The column I am using for this application is rated between pH 1 and 12. If anyone has any other mobile phases that would be easy to prepare and LC/MS compatible I would appreciate any ideas. If you think what I have will work I would appreciate hearing your feedback on that as well.

Thanks in advance.
for pH 11 you'll probably want to use ammonium bicarbonate
A. Carl Sanchez
A. Carl Sanchez,

Thank you for your response. I will certainly give that a try.

Uplc4me
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry