If you want to separate two compounds, and you run TLC a few times with different solvent systems, how do you determine from the TLC plate which is the best solvent system to separate the two compounds from each other using column chromatography?
Do I measure the Rf values of both compounds on each plate? I've read that an Rf of about 0.30 is good, but both spots can't have an rf of 0.30 otherwise they won't separate properly.
Or do I determine which plate gives the best Rf value for each compound. Say plate three gives the best Rf for compound A and plate one gives the best Rf for compound B. Then use the solvents system for plate three to elute compound A and then switch to the solvent system used for plate one to elute compound B?
What is it that you want to optimise ?
Do you want to get an acceptable separation in the least possible time ? Do you have lots of time and need a very clean separation ? Does column length figure in you optimum ?
When you say "column chromatography" do you mean HPLC or low pressure, and how will you be detecting your compounds when you go to the column stage ?
Peter