GC FID & TCD Precision and Linearity

Posted: Fri Jun 17, 2011 12:25 pm
by bob82
Hi,

I've been reading this site for a while but this is my first post.

Would anyone have suggestions in regards to a solution which I might be able to make up to check my 6890 on an interm bases. We currently have a service provider onsite yearly to PMOQ the system but I would be looking to do a 6 monthly internal check on the GC. I'm just wondering would anyone else do this type of check and if so what solutions and concentrations would you use? The GC configuration is a 6890 split/splitless inlet going to a FID and on the back split/splitless inlet going to a TCD. I was thinking would it be possible to use an Ethanol in IPA type solution or maybe a tetradecane and hexadecane in hexane as the precision and internal standard to check the system.

Any suggestions would be a great help,

Bob82

Re: GC FID & TCD Precision and Linearity

Posted: Fri Jun 17, 2011 2:49 pm
by Axar
I guess this might help you:
http://www.edqm.eu/medias/fichiers/Anne ... ipment.pdf
There is some methods for GC qualification

Re: GC FID & TCD Precision and Linearity

Posted: Fri Jun 17, 2011 4:35 pm
by Consumer Products Guy
Why not just run ChemStation automated system suitability on each set of calibration standards, check RSD of retention times, peak areas, tailing factor, and resolution? That's what we do for our cGMP assays.

You can use the same injections for quantitation as well.