some question about GC/MS from a begginer
Posted: Mon Mar 21, 2005 2:36 am
dear all,
I have just begun to work with MS/GC for my project, and unfortunatelly nobody from my lab group have worked before with MS/GC, so I am a little bit lost.
I am starting to work with this MS/GC and I having some problem (maybe there are obvious for you guys).
1. when I run the experiments I always have a background of water in the MS (arround 120000 in absorbance). I have checked for leaks and the result is tha I have a great peack for water
(from 70-200 % with respecth amu 69 in the tune mode) but the peak for air remains small (2-10 %). The pressure is arroun 1 10-5 torr ( I think this pressure is somethin high, but it is possible to run the experiment ??
). So it that a sign of a leak in my system???
2. if I use the scan mode, the cromatoghraph I obtain have several peaks some of then at high amu (199 and bigger). is that a problem with the mechanical pump oil (I have installed a mist trap) or is because my MS is dirty???
3. In the future I will have to work with high concentration of solvent and I don't want to burn out my detector. Could any one give me an estimation fo the maximum signal I can have in the detector and about how much time ???
I have a HP 5971A/5980 series II MS/GC
Thanks to all for any solution to my problems, I really appreciate
I have just begun to work with MS/GC for my project, and unfortunatelly nobody from my lab group have worked before with MS/GC, so I am a little bit lost.
I am starting to work with this MS/GC and I having some problem (maybe there are obvious for you guys).
1. when I run the experiments I always have a background of water in the MS (arround 120000 in absorbance). I have checked for leaks and the result is tha I have a great peack for water
2. if I use the scan mode, the cromatoghraph I obtain have several peaks some of then at high amu (199 and bigger). is that a problem with the mechanical pump oil (I have installed a mist trap) or is because my MS is dirty???
3. In the future I will have to work with high concentration of solvent and I don't want to burn out my detector. Could any one give me an estimation fo the maximum signal I can have in the detector and about how much time ???
I have a HP 5971A/5980 series II MS/GC
Thanks to all for any solution to my problems, I really appreciate