Missing points in the Galaxie Calibration Curve

Posted: Thu Jun 16, 2011 1:58 pm
by TesoroSLC
I am using the Galaxie Data System with the Varian 3800 GC. When I try to set up a seven point calibration all that I see in the calibration curve is only one point - the last point in the level. Anyone know what needs to be done to display all seven points? :(

Re: Missing points in the Galaxie Calibration Curve

Posted: Mon Jun 20, 2011 10:29 am
by Spuzzin
Hi,

Sounds like you are using the "clear all points" setting when you're doing your calibration. This would explain why you are only seeing the most recent calibration point. Change "Clear all points" to "add" and see how it goes.

Rich

Re: Missing points in the Galaxie Calibration Curve

Posted: Thu Jul 21, 2011 11:18 am
by petroleum
HI,

are you clicking on Level 1???