by
Chroman » Fri Jan 28, 2011 7:15 pm
How old is your trap? How old is the column? Do you have another column you can try?
Lucky that you can desorb for <4 min. 524.2 says "about" 4 min. The lowest my auditors will allow is 3 min!
.
Is your column flow realy 20 cc/min.? When I was using a 60M X 0.25 column I was flowing 1.7 cc/min.
Both column and trap are three months old.
We are just starting our certification process so no auditor has seen it yet..
I already installed a new column because I noticed in the last standard I ran that all peaks that elute before the beginning of the final ramp are tailing big time.
20 mL/min is the total flow, column flow is 1.5 mL/min