how to test melamine
Posted: Tue Jun 14, 2011 3:57 pm
by vishnu
I am analyzing metformin as per Ph.Eur. for that the starting material is dicyandiamide (dcda).so i need to analyze dcda as it is the primary raw material. in that melamine (tricyantriamide) is an impurity. i tried some methods given by the vendors. but it is not giving the satisfactory result. i even tried by spiking melamine with the dcda. but it is not giving the correct result.
Can anyone please help me???
Re: how to test melamine
Posted: Tue Jun 14, 2011 4:13 pm
by DSP007
Posted: Tue Jun 14, 2011 5:17 pm
by Vlad Orlovsky
Hello Vishnu,
Please send me a request with your email and I will share our method for direct analysis of melamine in milk without any sample prep. It employs Primesep column and guard Our report is in English, and you don't need to translate it from Russian...even with Google
here is a link on method:
http://www.sielc.com/Compound-Melamine- ... amine.html
and on the set up:
http://www.sielc.com/upload/file/pdf/SI ... r_2004.pdf
Contact me if you have questions.
Posted: Wed Jun 15, 2011 2:25 am
by vishnu
thanks all for the reply,
you can send me the method to my mail id,
vshr82@gmail.com or to
vishnusank1@gmail.com.
I am not testing for milk products, i am testing for pharmaceuticals and i need to test in 218 nm in a cationic exchange column using 3 pH buffer as mobile phase.
what i need is a clear peak of melamine when i inject dcda with good resolution.
can anybody help???
Posted: Wed Jun 15, 2011 6:15 am
by DSP007
can anybody help???
Good day, Vishnu
Yes of course, but milk a more complicated object than the pharmaceutical substance.
The Chinese theme search for traces of melamine after the scandal of "baby milk food"
well developed, the work of the classic. From here you can take the chromatographic system.
Generally I do not understand why you are working on such a low wavelength, the melamine maximum of about 280. Check also other impurities?
Posted: Wed Jun 15, 2011 7:38 am
by vishnu
i fixed that wavelength because our vendor is using the same and he is getting the peaks, but i am not getting in the same method and same conditions that they specified. i dont know the reasons.
is there any specified method for that? it gives me a huge head ache. without finding the method to test this melamine in cyanoguanidine, i cant validate anything. so please help me
Posted: Wed Jun 15, 2011 4:26 pm
by Markus Laeubli, Metrohm
We are using a cation column and 240 nm without any problem (see
Application Note U-42). Using lower wavelength might give more interferences but should work as well.
Posted: Fri Jun 17, 2011 1:53 am
by vishnu
Thanks for the reply.
But i am using the ammonium dihydrogen phosphate as mobile phase at 3 pH at 218 nm.
Is there any method related to that for determination of melamine in Dicyandiamide? because i cant change the mobile phase for this determination. or otherwise there may be a problem in quantifying DCDA.
Posted: Fri Jun 17, 2011 4:49 pm
by vishnu
Dear all,
i injected melamine spiked with dcda, it gave a peak. but it is so near to the dcda peak. so i cant quantify that.
Can i improve the resolution by doing something?
I am using whatman 10 scx column using ammonium dihydrogen phosphate in 3 pH as mobile phase.
can anybody help?
THANKS