- Posts: 4
- Joined: Sun Jun 05, 2011 11:01 pm

We are trying to set up a GC/FID but noticed a number of problems (drifting retention time, ghost peaks, FID areas were not reproducible). These chromatograms were made yesterday after cleaning the injection port, liner and insert and conditioning at 275C overnight. Also we added glass wool to the liner. The chromatograms, from top downward, are n-Octane, Ethylbenzene and o-Xylene, each sample has been injected for two runs to verify its repeatability, therefore the top two are n-Octane, middle two are Ethylbenzene and the lower two are o-Xylene. They are 1 microL injections, without solvent.
The machine is a HP 5890 series II GC/FID with HP autosampler 7673A, running with a Petrocol DH Capillary Column by Supelco with 100m*0.25mm*0.5micron film thickness. The method being used is injection port 220C, FID detector 220C, temperature ramped at 5C/min from 50C to 250C then hold for 10 minutes, with programmed syringe twice pre-washing, twice post-washing and 3 times pump. Carrier gas is Helium, FID is using Air and H2.
We measured the split vent and purge vent flow rate to be 104ml/min and 3ml/min, so we are assuming the split ratio to be 100:1 because of the description in the manual, or what other ways are there for us to judge the split ratio? This machine goes with a manual gas controller.
So the major problem is what you can see in the chromatograms, no effective signals(first one), drifting the retention time(3 and 4), suspicious remnant peak from the previous run(5 and 6),and unsteady peak area between runs.
Please help us, any comments and suggestions will be greatly appreciated!