I'm running a HILIC method on an Agilent 1100. I'm using a Phenomenex Kinetex 50x4.6mm 2.6-micron HILIC column with 86:14 A:B (A = 100% Acetonitrile with 0.1% TFA, B = 100% Millipore water with 0.1% TFA) at 1 mL/min and 30 degrees C, monitoring 257 nm, with a 10 uL injection volume.
Samples are dissolved in 9:1 Acetonitrile:Water with 0.1% TFA. My analytes are Paraquat and Diquat.
Diquat elutes first, and is non-target in this instance, but is included in my standard mix. Paraquat, my target, elutes second. 257 nm is the UV max for Paraquat.
My problem is this: My retention times are shifting earlier with increasing sample concentration. The first chromatogram below shows a 0.02 ug/mL standard overlaid with a 0.5 ug/mL standard. The second chromatogram shows all my standards overlaid, which range from 0.02 ug/mL to 1 ug/mL.
I am having a hard time accepting that I am overloading the column at these concentrations, but that's what it seems like is happening. Am I missing something here?