phosphate solubility contributing to my noisy blank?
Posted: Sat Jun 11, 2011 3:00 pm
I've been troubleshooting a rather noisy blank for almost two weeks now, basically large peaks ranging in the 10K area count. This problem is apparent regardless of system used. I'm not allowed to change the organic modifier. I think I have the problem identified, but would like to have some input if what I am seeing sounds correct. This is a HPLC method being run on a agilent 1100 (binary pump),
mobile phase A is 5mM (total) phosphate buffer, pH7 (prepared by combining 2mM monobasic and 3mM dibasic potassium salts)
mobile phase B is 100% ACN.
Sample diluent is 80% ACN/20%water (v/v).
Injection vol is 5ul
The gradient is linear from 5%B to 75%B in 13 minutes.
Peak of interest elutes at ~12 minutes (~70% organic in the gradient).
When I replace the mobile phase A with 100% water, the baseline looks fine, so I thought we had a contaminate in one of the salts or in the water. We've selectively changed one variable at time ,either repreparing the salt with a new bottle or using bottled water. The baseline always reverts to a similar level of noise when the buffer is used as Mobile phase A.
Then, I read an article from Schellinger and Carr (LCGC Northamerica, vol22, number 6, June 2004) regarding the solubility of various phosphates in 3 different organic modifiers. One conclusion reached in the paper was that phosphate buffer (at pH7) has 5mM solubility at 80% ACN. Hmmm...my diluent does not have any buffer in it, BUT it does have 80% ACN. I tried to add (dropwise) some of the buffer in a small beaker of the sample solvent and I do NOT see any precipitation...BUT I am wondering that the combination of the high organic in the diluent PLUS the organic being contributed in the gradient could be perturbing the salt enough to make the baseline noisy.
I then prepared two additional diluents: 50:50 ACN:water and 30:70 ACN: water, keeping the 5mM PO4 as mobile phase A. The purpose of these was to in a stepwise fashion, increase the water content in the diluent (blank) to test my hypothesis that the phosphate needs more water. There is a reduction in noise as I increase the water content in the diluent, but it is not completely gone.
Hypothesis: There is no contaminate causing the noise, but there is an apparent threshold I am approaching that even at 5mM PO4, which is causing the salt to "crash out" and perturb the baseline with a combination of 80% organic (or as low as 30% organic) in the diluent PLUS the organic being contributed in the gradient (~70%) . As I type this, perhaps if I am correct in this hypothesis, I could try to premix the mobile phase B to 80:20 ACN:water to additionally increase the water content at the point of mixing in the pump.
Question(s): What are the limits in terms of organic content people typically use with potassium phosphate salts? My next step was to try to use ammonium phosphate which according to the article, has greater solubility. Would the use of ammonium phosphate adequately buffer ~pH7. Would there be any significant change in chemistry of the separation going from a PO4 to a NH4 salt?
Thanks in advance for your assistance. Comments to my thoughts would be greatly appreciated.