waters alliance 2695 injection issue

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
After an injection is made, the volume in the sample vial INCREASES by about 25%. I changed needle, frits and filters.
The orientation of the needle hole is correct. Could needle wash be leaking into vial?

any thought?

thanks
There are 3 valves in the AS portion of the 2695. One (or more) of them is malfunctioning (probably valve 3 if memory serves). Call Waters service and they will help you diagnose it.
Valve 3 is usually the one that goes bad. The other two spool valves almost never fail (I've seen one go bad on a 717 in 20 years - they're the same on a 2695). With that said, you mentioned that you changed the needle, frits, and filters. Did you change the the seal pack?

CJ
http://the-ghetto-chromatographer.blogspot.com/
all set now, thanks for the replies, I got the seal pack adjustment to pass using IPA. (way better than Methanol)
V1 had a seal that had slightly popped out.
4 posts Page 1 of 1

Who is online

In total there are 17 users online :: 3 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], anu990 and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry