Response factors and Y-intercepts
Posted: Thu Jun 09, 2011 2:16 am
Hi all, I am relatively new to HPLC but have been assigned to a very massive method development project- yey. I've been having trouble establishing response factors for external standards. The problem is, when my two standard curves have non-zero y-intercepts (which is always), I cannot establish a Response factor between the two substances except at specific concentrations, as the Response factor is specific to that concentration. This is because the Area/Concentration ratio varies based on the concentration. I am calculating the RF via (AreaA/ConcentrationA)/(AreaB/ConcentrationB)=RF. Am I misunderstanding the use of response factors or how to calculate them? Secondly, shouldn't the y-intercept be zero; and the fact that it isn't an indication of a poor run (ie. overloaded etc etc) or non-linearity?
Some examples:
If analyte A's standard curve is A=2x+1. At x=1 A/x=3/1, At x=2 A/x=5/2.
If analyte B's standard curve is A=3x+4. At x=1 A/x=7/1, At x=2 A/x=10/2.
At concentration 1, a response factor of 3/7
At concentration 2, a response factor of 2.5/5
Thanks!
Wes
Some examples:
Thanks!
Wes