Do you think can I measure Carbon Monoxide in Shimadzu 2014? FID detector and column is DB-624 (Agilent)
Classic column of GC.
Where can I find information about the method?

Thanks a lot.
You do not have the correct detector or column.

Usually an FID is not sensitive to CO. Using a DB-624 means that CO is not separated from many gases which could interfere.

You should be using a mole sieve column and either a TCD, or a Helium ionization detector, or a FID with hydrogen carrier and a methanizer column before the detector.

The application will depend upon the matrix you are injecting.

Do a little research from process control analyzer companies or hire a professional to help you.

Good luck,

Rod
As Chromatographer 1 says, matrix dependant.

However we are doing ppm level CO in H2 using ppq pack column then onto a methaniser before FID detection.

What is the CO in?
You can use FID with methanizer or TCD detector. I think you can use a Porapak Q Plot column for CO analysis.
4 posts Page 1 of 1

Who is online

In total there are 8 users online :: 2 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry