- Posts: 30
- Joined: Tue Dec 21, 2010 3:39 am
I managed to separated my eight interest compounds using standard. However, the compounds show different intensity in fluorescence and UV even the conc. are same. This makes the peaks not fully separate. So, I reduced the conc. of one compound in the standard mixture according to roughly ratio that can be found in sample. The compound usually present in low amount or high when the other two side peaks absent based on literature review. Now, the peaks are separated well.
My questions is
1) Is what I did correct and valid for next validation step?
2) Should I do the validation step by injecting the mixture of standards or by individually the compounds to determine linearity, range, LOD,LOQ and etc?
Thank You.