We are using a 6890/5973 in EI mode. We have a PTV installed, and when it works, it is great and when it doesn't, it is awful. Anyway, we had service yesterday on the PTV (using way too much liquid nitrogen--turned out to be a loose nut somewhere in between the GC and MS). My tech is trying to tune the instrument, and we are seeing 32 and 28, and they pulse every 20 scans or so. They are in the 4:1 ratio I'd expect for an air leak. m/z 32 pulses to about 35-40% of the 69 from PFTBA, and then goes right back down--the pulse lasts about 2-3 scans. I've done troubleshooting on air leaks before, but I have never seen one pulse like this. Does anyone have ideas where to look? This happened after service, so I'm inclined to check things myself at the moment.
Thanks in advance!
Michele