GC-MS RIC versus Quant. Ion

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I am running a headspace GC-MS method similar to USP <467> for the analysis of volatile solvents in a polymer. When processing the data I get very different results depending on whether I use the RIC (reconstructed ion current) versus the Quant. Ion (m/z=73 for Methylethylketone). I am using external calibration with 3 calibrants. The results are 73 ppm MEK with RIC and 205 ppm with Quan. Ion m/z=73. Does anyone know how to explain this? Is there any guidelines for using RIC versus Quan. Ion for quantitative results?

Thanks in advance.

Regards,

Chris

In the RIC mode the detector acts as a universal detector. It is possible that something is co-eluting with your analyte which will affect the results. In the SIM mode it is possible that this co-eluting substance becomes invisible to the detector, so you may get different results.

One more thing- I don't understand why you are looking at 73- is this a typing error? The MW of your analyte is 72, which means you would be looking at the carbon 13 isotope peak of the analyte!!!! :shock:
2 posts Page 1 of 1

Who is online

In total there are 43 users online :: 1 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry