Currently, i had face the problems of all the peaks tend to elute at early of chromatogram (50 min). Previously, i had run the same sample with the same condition, the peaks were separated well and major peak is eluted at 18 min. My analyst is polar compounds.
Column : C-18
Solvent (A): Water
Solvent (B): Acetonitrile
Gradient elution (start with 100 % water)
I had noticed that the pump pressure is fluctuated (30 bar -> 24 -> 17 -> 19 -> 20 -> 30 bar).
There was air bubble occurred at the tubing after 2 solvent were mixed. ( I had degas the solvent manually and the degasser of the HPLC was spoilt.)
Besides, i noticed that the pressure during condition of column (water) was lower than flushing of column (33 bar)
are these conditions cause all my peaks shifted?
Really appreciate for all of your opinion.