Been trying to develop a method to separate the enantiomers of 1-phenylethanol. If I run the same method as that in the test chromatogram that came with the column, my separation does not appear to be as great (close, but theirs has a distinct baseline in-between and mine seems to have a little more tailing) - first and second chromatograms in image below. Now if I alter the method to fully separate the two peaks, I end up with more tailing (third chromatogram). Any solution to this or I should just accept this tailing as a fact of life?

This is a brand new Astec/Supelco B-DM column (30 m x 0.25 mm x 0.12 um), equipped with a guard column. New inlet liner, septum, checked for leaks and already re-installed column. Tried *many* different ramps/flows (all with D=I=250 degC and split ratios varying between 30:1 to 100:1 (depending on my sample), a different sample solvent (peaks before those, like anisole and acetophenone, seem to come out more symmetrically).
Thanks!
Roxanne.