Analytes that coelute

Posted: Wed May 25, 2011 10:41 pm
by pmflanders
I'm working with 1,4-benzoquinone and the herbicides chloramben, aminopyralid, and clopyralid. I'm having problems with them eluting too close to the dead time and at the same time. I've tried a bondapak C18, an Atlantis T3, and u bondapak phenyl 25cm columns with 0.24% acetic acid and methanol as the mobile phase. I have the same problem with any ratio of buffer to methanol in isocratic mode (even down to 1% methanol) and any gradients I've tried. The herbicides are carboxylic acids, so I used acetic acid to try to keep them neutrally charged.

This instrument is UV-HPLC, but I also use LC/MS, so I tend to stick with volatile buffers.

Are there any other tricks that I've missed?

Re: Analytes that coelute

Posted: Wed May 25, 2011 11:16 pm
by tom jupille
How about ion-pair chromatography using tetrabutylammonium ?
or one of the mixed-mode columns ?

Re: Analytes that coelute

Posted: Thu May 26, 2011 6:51 am
by Gerhard Kratz
Just a little google and you will find several applications for herbizides on HILIC columns.

Re: Analytes that coelute

Posted: Thu May 26, 2011 3:23 pm
by Vlad Orlovsky
This is a very good case for mixed-mode chromatography where you can adjust retention by changing 3 variables: amount of organic, buffer concentration and buffer pH. This approach is clearly shown in this graph:
http://www.sielc.com/Technology_2D_Properties.html

and here are few examples for separation of closely related isomers:
http://www.sielc.com/Compound-O-Toluidine.html
http://www.sielc.com/Compound-2-Aminobenzoic-Acid.html
http://www.sielc.com/Compound-2-4-Dihyd ... -Acid.html

if you can provide me with sample we will develop a method for your free of charge (usually takes us couple of hours)

Re: Analytes that coelute

Posted: Fri May 27, 2011 11:51 pm
by Camisotro
I'm working with 1,4-benzoquinone and the herbicides chloramben, aminopyralid, and clopyralid. I'm having problems with them eluting too close to the dead time and at the same time. I've tried a bondapak C18, an Atlantis T3, and u bondapak phenyl 25cm columns with 0.24% acetic acid and methanol as the mobile phase. I have the same problem with any ratio of buffer to methanol in isocratic mode (even down to 1% methanol) and any gradients I've tried. The herbicides are carboxylic acids, so I used acetic acid to try to keep them neutrally charged.

This instrument is UV-HPLC, but I also use LC/MS, so I tend to stick with volatile buffers.

Are there any other tricks that I've missed?
Looking at the structures of these herbicides - what form do these take when neutrally charged? Chloramben and Aminopyralid both have a free primary amine group (quite basic) and a carboxylic acid (quite acidic). Is it reasonable to assume they are both zwitterions? Possibly Clopyralid as well.

Could this account for why they just refuse to retain on a C18 column even under conditions when they should be neutral molecules? This all points to the conclusion that a different retention mechanism on a different column would help you out here. Your analytes are inescapably polar, so you should be taking advantage of that. :)

It's also worth investigating if other mobile phase additives would have better ion-pairing effects, to help with retention here.

Re: Analytes that coelute

Posted: Mon May 30, 2011 12:49 pm
by Bintang
We do not have any application notes for these compounds but for some fluorinated compounds and amino acids that may give som ideas on where to start.

http://www.sequant.com/files/documents/ ... _Dimer.pdf

http://www.sequant.com/files/documents/ ... _Dimer.pdf

http://www.sequant.com/files/documents/ ... lanine.pdf