Have been getting a lot of HPLC knowledge thru the Forum, so first let me thank all the input.
I am developing a HPLC method for WS vitamins.
I manage to separate all 9 vitamins (B1,B2,B3,B5,B6,B12,Biotin,Folic acid and Ascorbic acid).
My replicates for standards are stable and no shifts at all. But when i inject the samples, the replicate injection will have all peaks shift late. And when go on to the next sample it gone back to the normal RT.
My HPLC condition: 5uL injection volume. 0.6ml/min. room temp. gradient mobile phase of phosphate buffer and acetonitrile. (max at 50% acetonitrile). I have a 10 min condition at the starting mobile phase concentration before injections. So I don't think it's the gradient change that cause the peak shifts (and I assume it only cause peak shift early anyway).
Thank you for the help.
Regards.